hello all,
I am working with a sample contains saponins from a plant extract. The sample solvent contains plant extract in Phospate buffer (pH 4.5). The mobile phase consists of gradient starts from (A,D 90-10) to (A,D 55-45) in 30 mints. The organic phase (D) is acetonitrile contains 0.05% TFA. The chromatogram contains some peaks having peak tailing. I am using C4 (300 A, 25cm, 4.6 id) column reverse phase at 25c.
1. How can i improve the peak tailing? Is the tailing normal in phytochemical analysis.
2. How can i improve the resolution of peaks? Is the weak resolution is normal in phytochemical analysis?

Thanks for your time.

Regards
C18 is good but I would first use a 100Angstrom (after bonding) pore size material and I would add a little bit of Triethylamine. A HILIC column I would also try. When you want later an MS as detector please replace the phosphate buffer with a volatile buffer system. Use always a guard column. Good luck.
Gerhard Kratz, Kratz_Gerhard@web.de
thank you for suggestions
3 posts Page 1 of 1

Who is online

In total there are 16 users online :: 0 registered, 0 hidden and 16 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 16 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry