To expand further here is how the dissolution procedure was made.
Initially, the vessel contains 500ml of Disso Medium A. After 1hour, 6ml of sample solution was withdrawn and replaced with exactly 6ml of disso medium A. Adding disso Medium B was made.
After 4, 8 and 24 hours, 6ml of sample was withdrawn and for every withdrawal, 6ml of disso medium B was being added.
1. For standard preparation, various concentrations were prepared at a 5%, 20%, 50%, 80%, 100% and 120% solution. Linearity curve was established by plotting each percentage (on the y-axis) and average percent area (for the x-axis).
2. As per normal way of establishing linearity curve, concentration of standard (is on the x) while peak area (is on the y).
The reference of the procedure was using Method 1 instead and generated a formula something like the following:
%dissolved = (peakareaofsample)(slope) + y-intercept + (6ml * X/volume)
where:
peakareaofsample: peak area of sample from the withdrawn solution
X: Percent Amount dissolved from the previous interval, corresponds to 0.0 for 1 hour
volume: corresponds to 500ml for the 1 hour and 1000ml for the remaining of the test.
I followed the above formula and compared the result generated by using Method 2 way of calculating as follows:
SampleConcentration = (samplearea - intercept) / slope
percent dissolved = (sampleconcentration*vesselvolume*100)/LabelClaim
After evaluating the results, I find both of them was calculating same values but my colleague doesn't want to accept Method 1 way of calculation because it somehow violates the way of creating linearity plot and that is concentration (as the x) and peak area (as the y). Just wondering, if Method 1 is not correct, How come that every time I'm comparing the results generated from both Methods, they are just the same? Do I have to accept the fact that Method 1 is not right? Thanks a lot for responses.