I’m looking for some advice to improve my chromatography. As a warning statement, I’m fairly new to HPLC, and I’m working with a compound that is difficult to separate on prep scale.
I’m separating a compound from a mixture very similar derivatives/impurities – appox 80% pure originally -- using a normal-phase prep column (Rx-SIL). Sample is dissolved in DMSO. Usually manually inject 1mL of concentrated sample in a 2mL loop. Mobile phase is mostly ethanol, first 100% then some gradient action.
I’ve done a lot of method development, and the chromatography is pretty ugly. Very poor resolution. Still, analyzing fractions by analytical HPLC (which has great resolution), I am usually able to collect a few fractions that are 95% pure, which is fine for my purposes.
I need a high volume of high purity compound and am continuously running performing prep runs. Unfortunately, the chromatography is not consistent. I will get runs with numerous high purity fractions (95%+), and other runs where the highest purity fraction is 92%. Shape and retention time varies slightly.
I equilibrate at 100% EtOH for 20 or 30 mins. To wash the column, I inject 2mL DMSO and run 70:30 ethanol:water w/ 0.1% acetic acid for about 60 mins, then back to 100% ethanol. I equilibrate and go to the next run.
As someone relatively new to HPLC, just looking for some brainstorming for what could be causing the inconsistency or anything I can do to encourage consistency.
One thought I had is precipitation in the column due to high concentration of sample injection / low solubility in ethanol. On top of this, my wash method is probably not ideal, so advice there would be great.