Inconsistency in Preparative HPLC
Posted: Thu Jul 10, 2014 5:51 pm
Hello,
I’m looking for some advice to improve my chromatography. As a warning statement, I’m fairly new to HPLC, and I’m working with a compound that is difficult to separate on prep scale.
I’m separating a compound from a mixture very similar derivatives/impurities – appox 80% pure originally -- using a normal-phase prep column (Rx-SIL). Sample is dissolved in DMSO. Usually manually inject 1mL of concentrated sample in a 2mL loop. Mobile phase is mostly ethanol, first 100% then some gradient action.
I’ve done a lot of method development, and the chromatography is pretty ugly. Very poor resolution. Still, analyzing fractions by analytical HPLC (which has great resolution), I am usually able to collect a few fractions that are 95% pure, which is fine for my purposes.
I need a high volume of high purity compound and am continuously running performing prep runs. Unfortunately, the chromatography is not consistent. I will get runs with numerous high purity fractions (95%+), and other runs where the highest purity fraction is 92%. Shape and retention time varies slightly.
I equilibrate at 100% EtOH for 20 or 30 mins. To wash the column, I inject 2mL DMSO and run 70:30 ethanol:water w/ 0.1% acetic acid for about 60 mins, then back to 100% ethanol. I equilibrate and go to the next run.
As someone relatively new to HPLC, just looking for some brainstorming for what could be causing the inconsistency or anything I can do to encourage consistency.
One thought I had is precipitation in the column due to high concentration of sample injection / low solubility in ethanol. On top of this, my wash method is probably not ideal, so advice there would be great.
I’m looking for some advice to improve my chromatography. As a warning statement, I’m fairly new to HPLC, and I’m working with a compound that is difficult to separate on prep scale.
I’m separating a compound from a mixture very similar derivatives/impurities – appox 80% pure originally -- using a normal-phase prep column (Rx-SIL). Sample is dissolved in DMSO. Usually manually inject 1mL of concentrated sample in a 2mL loop. Mobile phase is mostly ethanol, first 100% then some gradient action.
I’ve done a lot of method development, and the chromatography is pretty ugly. Very poor resolution. Still, analyzing fractions by analytical HPLC (which has great resolution), I am usually able to collect a few fractions that are 95% pure, which is fine for my purposes.
I need a high volume of high purity compound and am continuously running performing prep runs. Unfortunately, the chromatography is not consistent. I will get runs with numerous high purity fractions (95%+), and other runs where the highest purity fraction is 92%. Shape and retention time varies slightly.
I equilibrate at 100% EtOH for 20 or 30 mins. To wash the column, I inject 2mL DMSO and run 70:30 ethanol:water w/ 0.1% acetic acid for about 60 mins, then back to 100% ethanol. I equilibrate and go to the next run.
As someone relatively new to HPLC, just looking for some brainstorming for what could be causing the inconsistency or anything I can do to encourage consistency.
One thought I had is precipitation in the column due to high concentration of sample injection / low solubility in ethanol. On top of this, my wash method is probably not ideal, so advice there would be great.