- Posts: 14
- Joined: Tue Dec 27, 2011 3:47 pm
Purge 11min.
Inlet Head Pressure 12.33psi
Dry Purge 1min.
Inlet Temperature 250C
Desorb 3min. 255C
Bake 10min. 270C
Mode Split
Split Ratio 65/1
Valve Oven 150C
Total Flow 42.9mL/min
Transfer Line 150C
Column RTX-VMS
20m X 0.18mm I.D.
1.0um Film Thickness
Oven Temp. Program
40C-100C at 10C/min
100-250C at 25C/min hold for 2min.
Column flow rate 0.6mL/min.
Mode Constant Flow
Gas Saver 150mL/min. @1.50min.
The problem is that as a run goes on I see less and less internal standard. Monday morning, after the mass spec hasn’t been used all weekend, is the highest the internal standard will be all week. For example, Monday morning my Fluorobenzene will be 700,000 area counts. 20 samples later, fluorobenzene will be 400,000 area counts. Tuesday morning, fluorbenzene will be 550,000 area counts, down to 350,000 at the end of the run. By Friday, if the equipment has been used all week, the starting sample will be almost half of what it was Monday morning. Also, once I have lost that much area, even after a weekend, it never goes back to where it was. After the week used in the example, I would be lucky to see 600,000 area counts on Monday morning.
The transfer line was replaced in September, trap was replaced in November.
I understand this is probably a water carryover situation, but don't know how to fix it.