Internal Standard recoveries dropping after calibration
Posted: Tue Jul 08, 2014 8:39 pm
Hello.
I'm running an HP 5890 GC with an HP 6890 MSD attached. I've just cleaned the source. I target tuned the MS after letting it set overnight.
Also, I've recently replaced the IS/SURR valve. My issue is this. I calibrated the instrument - the calibration looked fine maybe 5 compounds on linear curve fit but all other analytes are on average response. Looks lovely. Next morning I prep and run an ICV - secondary source calibration check - and my internal standards have all dropped to 60% so all compounds are failing high. I've run several rinses and the IS appears to be steady now. Should I simply recalibrate or is there an obvious underlying issue that I'm not aware of. Did the MS just need to stabilize a little longer? Thanks for your help.
I'm running an HP 5890 GC with an HP 6890 MSD attached. I've just cleaned the source. I target tuned the MS after letting it set overnight.
Also, I've recently replaced the IS/SURR valve. My issue is this. I calibrated the instrument - the calibration looked fine maybe 5 compounds on linear curve fit but all other analytes are on average response. Looks lovely. Next morning I prep and run an ICV - secondary source calibration check - and my internal standards have all dropped to 60% so all compounds are failing high. I've run several rinses and the IS appears to be steady now. Should I simply recalibrate or is there an obvious underlying issue that I'm not aware of. Did the MS just need to stabilize a little longer? Thanks for your help.