Reccomended P&T system?

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

5 posts Page 1 of 1
I am looking to replace an old Archon/Stratum combo with a new system. Our last two systems were Atomx, but we have not been able to run a method equivalent to USEPA 8260 (VOCs) satisfactorily. We do have a reduced purge time, but this was not an issue with the Archon systems.

We run all samples (waters or methanolic soil extracts in water) on soil mode (purged on the autosampler).

i am considering an EST CEnturion/Evolution but i am having trouble finding a supplier in Australia (can anyone help?).

I have also received some info on a CDS 7400, 7000 system

Any reccomendations?
----suffers separation anxiety----
I am looking to replace an old Archon/Stratum combo with a new system. Our last two systems were Atomx, but we have not been able to run a method equivalent to USEPA 8260 (VOCs) satisfactorily. We do have a reduced purge time, but this was not an issue with the Archon systems.

We run all samples (waters or methanolic soil extracts in water) on soil mode (purged on the autosampler).

i am considering an EST CEnturion/Evolution but i am having trouble finding a supplier in Australia (can anyone help?).

I have also received some info on a CDS 7400, 7000 system

Any reccomendations?
I have been running the Centurion/Evolution combination and really like them. Our first one was the third Centurion made that could run soils, and the software was a little quirky at first but EST did a good job working out the bugs quickly and keeping us updated. Now that unit can run non stop from Monday morning until sometime on the weekend without a hitch.(once a quarter we receive about 300 drinking water samples in a week and it runs right through them no problem)
The past is there to guide us into the future, not to dwell in.
Thanks!

Do you ever have issues with stuck vials?
----suffers separation anxiety----
With the Centurion we don't have any vial problems since it does not move the vials unless it is for soils. We did have some problems with it removing the vial from the soil sparge position when we got the first one, but EST actually modified the design because of our feedback. Seems the vials we used were slightly larger(fraction of a mm)than what they used, and if they were wet it would form a suction after being placed into the cup, so they added small grooves to break the suction effect :)
The past is there to guide us into the future, not to dwell in.
Thanks for the input James!! I am VP at EST Analytical and would be glad to get you any information you might need. Please feel free to reach out to me at lpyron@estanalytical.com.
5 posts Page 1 of 1

Who is online

In total there are 21 users online :: 1 registered, 0 hidden and 20 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Majestic-12 [Bot] and 20 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry