Ion pairing

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

7 posts Page 1 of 1
Can ion pair reagents like OSA be used at pH 5-6? Thanks
Yes.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
Thanks Tom. What would be the max concentration recommended for UPLC systems?
I would say 0,1%, but please keep in mind that with ion pairing agents you do a modification of the stationary phase, irreversible.
Gerhard Kratz, Kratz_Gerhard@web.de
It's complicated! I wouldn't go much above 0.1%, but it depends on the % organic in the mobile phase as well. You certainly want to make sure you stay below the critical micelle concentration (CMC).

In our Advanced Method Development course (shameless plug: http://lcresources.com/training/tramd.html), we describe a systematic approach that basically looks like this:
Assuming a mixture of neutral and ionized analytes, first ignore the ionics and adjust the organic to give reasonably high retention of the neutrals (k' values between about 5 and 15 if possible). The ions will probably be unretained under these conditions.
Then do a series of runs with increasing ion pair concentration, going up in geometric progression (e.g., 1mM, 2 mM, 4 mM, 8mM, . . . or 1mM, 3mM, 10mM, . . . ) to "walk" the ions into the same retention range.

[edit addition] If you have a copy of the Snyder/Kirkland/Dolan "Introduction to Modern Liquid Chromatography" book, there's a table of recommended maximum concentration of different ion-pair reagents as a function of % organic on page 341.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
Thanks to all for the answers. I have been running at 50% ACN or 65% MeOH isocratic and would like to transition to the same mobile phase containing ion-pair reagent, could the organic content be a problem? Thank you.
Thanks to all for the answers. I have been running at 50% ACN or 65% MeOH isocratic and would like to transition to the same mobile phase containing ion-pair reagent, could the organic content be a problem? Thank you.
Make up at that concentration, see if anything precipitates.
7 posts Page 1 of 1

Who is online

In total there are 33 users online :: 0 registered, 0 hidden and 33 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 33 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry