Mass 31 when PFTBA is OPEN

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

2 posts Page 1 of 1
Hello All,

I have a 6890/5973 set-up. When in the Manual Tune is in Profile mode looking at peaks 69, 28, and 32, there is a peak at 31 that is between the abundances of the 28 and 32 - about 4,000 for the N2 as a reference. This peak at 31 is only present when the PFTBA is open and completely disappears when the vial is closed. Is there an easy fix for this? Is this something I should be more concerned about?

Regards,

m.finnBMBM
Hello All,

I have a 6890/5973 set-up. When in the Manual Tune is in Profile mode looking at peaks 69, 28, and 32, there is a peak at 31 that is between the abundances of the 28 and 32 - about 4,000 for the N2 as a reference. This peak at 31 is only present when the PFTBA is open and completely disappears when the vial is closed. Is there an easy fix for this? Is this something I should be more concerned about?

Regards,

m.finnBMBM
Was the cal gas vial ever rinsed with methanol? The mass you see from methanol is normally 31.
The past is there to guide us into the future, not to dwell in.
2 posts Page 1 of 1

Who is online

In total there are 30 users online :: 1 registered, 0 hidden and 29 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Bing [Bot] and 29 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry