by
NO89 » Sat Jul 05, 2014 9:03 am
Firstly, thank you all for taking your time to contribute to this discussion, I very much appreciate the time you have taken to respond.
I will try and answer / provide more information as I go along.
Right with regards to
Farooq, it is fronting! Remember that right after the 1. Control another one is injected (control 2) which is just a replicate. That control (2) doesn't exhibit the same fronting. Ergo a mobile phase composition inconsistancy is most probably the issue. More thorough review of the method parameters including the gradiet table is the way forward.
Control 2 inst exactly a replicate. Control 2 has a higher percentage of disialotransferrin, so in a sense control 1 is like a negative control where as control 2 is a positive control.
Farooq,
If possible, do a LC-MS of this shoulder, other simply alter the mobile phase (make it slightly weaker) so that the minor component is separated from this "tetra" peak.
Unfortunately here in the lab we do not have the equipment to perform MS, but this is something i will suggest to my superiors. As for weakening the buffer, this is a pre made buffer, it comes in already made, with no information regarding its pH or anything, however I will contact the suppliers and see if they can provide me with any further information.
The fronting / co eluting gets worse as the sample set progresses (the control 1 injection at the end of the analytical run is much worse that the control 1 injection at the beginning of the sample set).. please see attached image of the control 1 injection at the end of the analytical run:
Klaus I, could you please specify which type oif information you would like regarding the system? It uses a Quaternary solvent manager, with a pre column heater in which the analytes are detected via UV PDA detector. Current running Empower Pro software.
It looks like ”sample solvent/mobile phase” inconsistency – perhaps the control sample’s pH is higher or lower than the mobile phase’ and the rest of the samples.
You can also see that the peaks if the control sample elute slightly earlier than those of the test sample.
Maybe you’d like to post some more info for use in the further trouble-shooting process.
1. What is the mobile phase pH?
2. What is the pH of the rinsing eluents.
3. What is the pH of the sample solutions (both the control and the test)
4. How does the gradient table look like?
5. How long is each run?
6. How much volume do you load/inject?
After analysing yesterdays analytical run this fronting/co elution issue is now affecting my control 2, but this could be due to carry over from control one.
I do not know what the pH for the mobile phase is or the controls as this is not disclosed by the company that provides the buffer, but I will contact them to try and gather this information.
The gradient table is as follows : Start to 10 minutes is 100% Buffer A, from 10 to 10.1 minutes it is 60% buffer A and 40% buffer B. from 10.1 minutes to 14 minutes it is 100% buffer B. from 14.1 to 20 minutes it is buffer C.
The flow rate is 0.8 ml/min which buffer A runs through and 1.5ml/min when the other buffers are running through.
Each run is 20 minutes long, however we only use the first ~10 minutes of data.
The sample volume that we inject is 250ul.
Hollow,
you've said the peak is eluted whilst A is running through and B&C are used for column cleaning. So in fact the separation is in isocratic mode?
Yes that would be correct, when the peaks are eluting this only happens when A is running through.
And what about the detection? Any possibility to spectroscopically (PDA,MS) check the front shoulder if there are any differencies to the peak apex?
Again, Unfortunately this is something we do not have readily available here, it is something I would have to bring up with my superior.
Again, thank you all for your input. I will not be back in lab now until Tuesday, so apologies in advance if i can not respond to any more of you question until then.