GC789 with G1888 in empower

Discussions about GC and other "gas phase" separation techniques.

2 posts Page 1 of 1
Hello,

I'm trying to do a run with sequence with an GC7890 with G1888 under Empower 2. My ICOP can control G1888 (I can do it with GC6890 and G1888). My exact problem is when I did start my sequence, the G1888 do not follow my instrument method parameters. It return to last parameters saved in it. What is weird is if I do setup or monitoring all works fine and G1888 have the parameters I set in my instruments method, but I soon I start my sequence the parameters changes.

Someone have an Idea?

I know my english is not good so if I'm not clear please let me know
Hello to all,

I did found my problem :P On the G1888, the option "Chain Method" was activated. To be able to Empower to control it, I have to deactivate it. By the way to deactivate it you must put "0" (zero) as value. I had to searh to find out how to do it. I wish it could help other. :mrgreen:
2 posts Page 1 of 1

Who is online

In total there are 19 users online :: 0 registered, 0 hidden and 19 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 19 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry