β-cyclodextrin removal

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
Hello to all, im using an HPLC method with β-cyclodextrin in mobile phase. Im washing the column with water:ACN 80 :20 to remove this detergent from the column but the pressure in my column didnt fall at the starting level.Anyone know any other way to get rid off β-cyclodextrin from my column?
Thnx in advance
Try reverse-flushing the column.
-- Tom Jupille
LC Resources / Separation Science Associates