We're trying to develop a method to analyze acetone, butanol, ethanol, butyric acid and acetic acid. We are using a SRI 6810 GC splitless injection with 10 mL/min nitrogen carrier gas flow and uses Peaksimple software. The column we are using is a Restek MXT-WAX Polyethylene glycol nominal dimensions: 30.0 m x 530 µm x 1.00 µm.
At a different facility we were using an Agilent 6850C GC system with split injection ration of 50:1, variable nitrogen carrier gas flow and uses Chemstation software. The column there is an Agilent 19091F-115E HP-FFAP Polyethylene Glycol TPA capillary, nominal dimensions: 50.0 m x 320 µm x 0.50 µm. We were able to get good results using this setup.
With the SRI 6810 we are able to get good acetone, butanol and ethanol peaks, but the acetic and butyric acid proves to be more problematic. We also protonate the sample with 0.2 M HCl and we're worried about column deterioration as it is a splitless injection without an inlet liner. Are there any suggestions or different techniques that we could use to quantitatively analyze butyric and acetic acid?
Current method for SRI 6810:
The ramp oven temperature is initially set at 45 oC and held for 5.0 minutes. The temperature is then raised to 190 oC at rate of 30 oC/min and held for 2 minutes (ramp duration = 5.8 minutes). The total running time is 20 minutes. The nitrogen gas flow is fixed at 10 mL/min.
Current method for Agilent 6850C:
The ramp oven temperature is initially set at 80 oC and held for 0.5 minutes. The temperature is then raised to 200 oC at rate of 10 oC/min and (ramp duration = 7.5 minutes). The total running time is 20 minutes. The nitrogen gas flow starts from 0.8 ml/min and held for 8 minutes. It is then increased to 1.0 ml/min at an acceleration rate of 1.0 ml/min2 and held for 5 minutes. The flow rate is then raised to 1.3 ml/min with the same acceleration rate of 1.0 ml/min2. The total running time is 20 minutes.
Thanks in advance.
At a different facility we were using an Agilent 6850C GC system with split injection ration of 50:1, variable nitrogen carrier gas flow and uses Chemstation software. The column there is an Agilent 19091F-115E HP-FFAP Polyethylene Glycol TPA capillary, nominal dimensions: 50.0 m x 320 µm x 0.50 µm. We were able to get good results using this setup.
With the SRI 6810 we are able to get good acetone, butanol and ethanol peaks, but the acetic and butyric acid proves to be more problematic. We also protonate the sample with 0.2 M HCl and we're worried about column deterioration as it is a splitless injection without an inlet liner. Are there any suggestions or different techniques that we could use to quantitatively analyze butyric and acetic acid?
Current method for SRI 6810:
The ramp oven temperature is initially set at 45 oC and held for 5.0 minutes. The temperature is then raised to 190 oC at rate of 30 oC/min and held for 2 minutes (ramp duration = 5.8 minutes). The total running time is 20 minutes. The nitrogen gas flow is fixed at 10 mL/min.
Current method for Agilent 6850C:
The ramp oven temperature is initially set at 80 oC and held for 0.5 minutes. The temperature is then raised to 200 oC at rate of 10 oC/min and (ramp duration = 7.5 minutes). The total running time is 20 minutes. The nitrogen gas flow starts from 0.8 ml/min and held for 8 minutes. It is then increased to 1.0 ml/min at an acceleration rate of 1.0 ml/min2 and held for 5 minutes. The flow rate is then raised to 1.3 ml/min with the same acceleration rate of 1.0 ml/min2. The total running time is 20 minutes.
Thanks in advance.