We recently purchased a Thermo Scientific Ultimate 3000 HPLC system to quantitate biogenic amines and aminoacid neurotransmitters. We have problems programming Chromeleon 7.2 to make the autosampler perform derivatization of amino acids using OPA. That is to say we can easily make the autosampler take up 40ul OPA, transfer to 40ul sample, mix 4 times, wait 2 minutes and inject e.g. 15ul. But we get very poor reproducibility with our program (>10% variability). Performing the same derivatization manually using the regular injection routine of the autosampler brings the variability in the range of <1%.
Do any of you have experience in optimizing the chromeleon programming for mixing on the Dionex Ultimate3000 autosampler?
Here's what we have right now:
UdpInjectValve Load
UdpSyringeValve Needle
UdpDraw Air 5ul; GlobalSpeed; GlobalHeight
UdpDraw ReagentAVial: 40ul; GlobalSpeed; GlobalHeight
UdpDispense SampleVial 40ul; GlobalSpeed; GlobalHeight
UdpMix SampleVial; 40ul; 10; GlobalHeight; 4
UdpDispense Drain; 5ul; GlobalSpeed; GlobalHeight
UdpDraw Air; 5ul; GlobalSpeed; GlobalHeight
UdpMixWait 120
UdpDraw SampleVial; Sampler.volume: GlobalSpeed: GlobalHeight
UdpInjectMarker
UdpInjectValve Inject
UdpSyringeValve Waste
UdpMoveSyringeHome GlobalSpeed
UdpMixNeedleWash 50ul
Thanks in advance
Thomas