5975C MSD tuning stopped

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

3 posts Page 1 of 1
Hi all

We have a Agilent 5975C MSD. I tried to autotune and an error message appeared during tuning :

Tune action stopped:the valid range for ent lens offset is 0.00 to 64

anyone has any idea of the possible cause of this matter?

Thanks!
Reload your last good tune file.
Hello milenuj!

We have the same MSD and experienced the same error message. It might be difficult to find, but what the error says is, that the optimal voltage that would be applied to the entrance lens is higher than allowed. You can see this when tuning. The curve of the voltage ramp of the entrance lens does not have its maximum before 64V.

solution:

we took out the ion source and did two things:
1. the entrance lens itself has a contact which is connected to the cable applying the voltage. This contact was touching the outer metal surface of the ion source. Simple turning of the entrance lens was done.

2. Tech support told us, that there is this super small screw, which locks the inner insulator in the source body. This screw was said to possibly be responsible, when tightened too much. Try slightly loosening it.

This resolved the issues we were having with autotuning.
3 posts Page 1 of 1

Who is online

In total there are 18 users online :: 0 registered, 0 hidden and 18 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 18 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry