flushing the LCMSMS system in the end of the day

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

5 posts Page 1 of 1
Hi all
In the manual they advice to flush your LC/MS/MS system in the end of the day for 15 min with 50:50 methanol:water and they give values that should be set for most of the parameters in HESI mode. However, they did not state at which temperature the ion transfer tube should be set. Any idea about it.
Sorry for this basic question.
cheers
Hi,

What mass spectrometer do you use ?
As you're talking of HESI and Ion transfer tube, I guess you have a Thermo instrument, but what model ?
Hi
It is a quantiva.
For a Quantiva, I strongly advise you to let the ion transfer tube heated. Otherwise you will have a bad vacuum due to the huge size of the ion transfer tube.

I personally do that in all my instruments.
ok. Thanks very much :)
5 posts Page 1 of 1

Who is online

In total there are 49 users online :: 2 registered, 0 hidden and 47 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Amazon [Bot], Bing [Bot] and 47 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry