Advertisement

Degassing hexane

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

19 posts Page 1 of 2
Can hexane be degassed?? For how much time?? As its boiling temperature is 68 degree, it will get heated up.

If hot hexane passed through hplc tubing during purging for 1-2 min can it be damaging for the system??

As I am getting staright baseline with 100% Isopropanol and not with 100% hexane?? Can it be due to passage of hot hexane but isopropanol was also degassed but not passed hot while purging.
If you are degassing by purging the hexane will be cooled by its own evaporation, not heated up.

Peter
Peter Apps
Can hexane be degassed?? For how much time?? As its boiling temperature is 68 degree, it will get heated up.

If hot hexane passed through hplc tubing during purging for 1-2 min can it be damaging for the system??

As I am getting staright baseline with 100% Isopropanol and not with 100% hexane?? Can it be due to passage of hot hexane but isopropanol was also degassed but not passed hot while purging.
Have you really degassed hexane by heating it to boiling state and then pumped hot to the system ? :o

Although I'm not a cleaner here, but could this strange post be moved to Liquid Chromatography forum ?
If you are degassing by purging the hexane will be cooled by its own evaporation, not heated up.

Peter

I am degassing by using sonicator water bath and I degassed it for 15-20 min so it got heated up in the meantime.....its temperature was around 70-80 degree....and I passed it through system but not through column.....

Can baseline be not coming straight due to hexane.....as with IPA it is straight when passing through column....
Can hexane be degassed?? For how much time?? As its boiling temperature is 68 degree, it will get heated up.

If hot hexane passed through hplc tubing during purging for 1-2 min can it be damaging for the system??

As I am getting staright baseline with 100% Isopropanol and not with 100% hexane?? Can it be due to passage of hot hexane but isopropanol was also degassed but not passed hot while purging.
Have you really degassed hexane by heating it to boiling state and then pumped hot to the system ? :o

Although I'm not a cleaner here, but could this strange post be moved to Liquid Chromatography forum ?


It is moved...... :lol: :D
I am degassing by using sonicator water bath and I degassed it for 15-20 min so it got heated up in the meantime.....its temperature was around 70-80 degree....and I passed it through system but not through column.....
Then your column stays intact. BTW - I would rather degass it by filtration through PTFE membrane filter under vacuum than by sonication.
Can baseline be not coming straight due to hexane.....as with IPA it is straight when passing through column....
In my opinion contaminants in hexane may be the culprit.
It is moved...... :lol: :D
Only the administrator can move it :lol:
If you are degassing by purging the hexane will be cooled by its own evaporation, not heated up.

Peter

I am degassing by using sonicator water bath and I degassed it for 15-20 min so it got heated up in the meantime.....its temperature was around 70-80 degree since hexane boils at 69C it cannot have reached 70 - 80 C unless it was under pressure, which would have defeated the object of degassing it. Try degassing for a shorter period, if you are still seeing bubbles evolving after 15 min it is because the hexane is boiling, not because there is still dissolved air coming out of it.....and I passed it through system but not through column.....

Can baseline be not coming straight due to hexane.....as with IPA it is straight when passing through column....
Peter Apps
if you want to sonicate without heat, you could try degassing by other means. I am very unconvinced by sonication, especially in a water bath type sonicator.

Many years ago, a few of us tried it. We happened to have a portable oxygen measuring probe, and were getting fed up with an HPLC technician who sonicated their buffers in a water-bath in our rather small lab for hours every day, and was doing our ears.

We shook a bottle of water ferociously to get it saturated with air, and took an oxygen reading, getting 100% saturation. We sonicated for 30 minutes, which reduced the oxygen saturation of the water to 99%.

We then stirred under tap-pump-vacuum for about half a minute, and got it down to 10%.

Sonication is supposed to work by causing cavitation, into which the gases evaporate, to create proper bubbles; I think you need a far more ferocious sonication to create cavitation than most hplc labs have available. The only mechanism in action in a typical lab is a vague hope that any dissolved gasses will find the noise so irritating they'll run away of their own accord...
Degassing with Helium or a vacuum degasser will help you. Vacuum degasser please ask the manufacturer that the tubings are resistant to Hexane.
Good luck.
Gerhard Kratz, Kratz_Gerhard@web.de
Don't shoot to the sparrow with the cannonball. Start with simple, affordable solution :)
http://www.basinc.com/mans/LC_epsilon/P ... obile.html
Thankyou all for your suggestions...... I will try them and try to solve the problem........... :D :D
Are you sure, that your problem is caused by solved gas and not by hexane itself due to its own vapor pressure?

When I used hexane with a normal phase column I ever had the problem, that I got bubbles in my HPLC from hexane itself. The backpressure was just to low. I could avoid that problem by using a mixture of hexan and a second solvent (in my case I used Ethyl acetate).
I'm a fan of the helium sparge. I think it is a gold standard of sorts for degassing.
I'm a fan of the helium sparge. I think it is a gold standard of sorts for degassing.
Helium sparging of a volatile liquid like hexane will evaporate more hexane than goes through the column !

Peter
Peter Apps
19 posts Page 1 of 2

Who is online

In total there are 35 users online :: 1 registered, 0 hidden and 34 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot] and 34 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry