flushing the LCMSMS system in the end of the day

Posted: Fri Jun 20, 2014 3:19 pm
by RALS
Hi all
In the manual they advice to flush your LC/MS/MS system in the end of the day for 15 min with 50:50 methanol:water and they give values that should be set for most of the parameters in HESI mode. However, they did not state at which temperature the ion transfer tube should be set. Any idea about it.
Sorry for this basic question.
cheers

Re: flushing the LCMSMS system in the end of the day

Posted: Tue Jun 24, 2014 9:03 pm
by jenche
Hi,

What mass spectrometer do you use ?
As you're talking of HESI and Ion transfer tube, I guess you have a Thermo instrument, but what model ?

Re: flushing the LCMSMS system in the end of the day

Posted: Wed Jun 25, 2014 8:18 am
by RALS
Hi
It is a quantiva.

Re: flushing the LCMSMS system in the end of the day

Posted: Wed Jun 25, 2014 8:50 am
by jenche
For a Quantiva, I strongly advise you to let the ion transfer tube heated. Otherwise you will have a bad vacuum due to the huge size of the ion transfer tube.

I personally do that in all my instruments.

Re: flushing the LCMSMS system in the end of the day

Posted: Wed Jun 25, 2014 8:54 am
by RALS
ok. Thanks very much :)