We are using a agilent molecular sieve 5A column to separate permanent gases. It all of a sudden stopped separating the gases...they now appear as one large peak. I noticed that the traps for O2 and mositure were a little saturated so i changed the traps and ran again. The same issue occured. We then baked the column at 190 degress for 1.5 hour again ...same issue occured!. So we thought the column is screwed!...so i changed the column and ran again and the same issue is occuring...peaks are not separating...any ideas?
-
- Posts: 212
- Joined: Fri Dec 05, 2008 11:03 am
Also i forgot to add we are using a refinery gas analyzer...with valve switching etc
-
- Posts: 1773
- Joined: Fri Feb 09, 2007 10:33 pm
seamoro,
Have you verified you have flow through the column/detector? Have you verified the injection valve is turning? Have you verified the column switching valves are turning? Not just air moving but the actual valve moving.
Best regards,
AICMM
Have you verified you have flow through the column/detector? Have you verified the injection valve is turning? Have you verified the column switching valves are turning? Not just air moving but the actual valve moving.
Best regards,
AICMM
-
- Posts: 6
- Joined: Sun Jan 26, 2014 2:51 am
seamoro,
The bake-out temperature is insufficient. For 1/16” columns, bake out at 250C for four hours. For 1/8” columns, bake out at 300C for four hours. Make sure there is nothing else in the temperature zone that can be harmed by the bake-out temperature. If there is, then either the MS5A must be removed for bake-out, or the components with lower temperature limits must be removed for the mole sieve bake-out. In general, our company places mole sieves in separate temperature zones with other columns that can be heated to 250C or 300C.
Jed Sullenger
The bake-out temperature is insufficient. For 1/16” columns, bake out at 250C for four hours. For 1/8” columns, bake out at 300C for four hours. Make sure there is nothing else in the temperature zone that can be harmed by the bake-out temperature. If there is, then either the MS5A must be removed for bake-out, or the components with lower temperature limits must be removed for the mole sieve bake-out. In general, our company places mole sieves in separate temperature zones with other columns that can be heated to 250C or 300C.
Jed Sullenger
Custom Solutions Group LLC
1419 Avenue A, Katy, TX 77493
Telephone:
(O) 281-574-9999
(C) 832-540-9127
Fax: 707-215-0932
E-Mail: jed.sullenger@customsolutionsgroup.org
1419 Avenue A, Katy, TX 77493
Telephone:
(O) 281-574-9999
(C) 832-540-9127
Fax: 707-215-0932
E-Mail: jed.sullenger@customsolutionsgroup.org
-
- Posts: 6
- Joined: Tue Jun 03, 2014 6:55 pm
Are you using MS-5A column with back flush ?. Looks like your valve timings have been changed, either your sampling valve timing is off in the event table or has a wrong value. Try increasing the sampling valve timing and see. If you have a valve diagram with you then it is easy to trouble shoot the problem. Also the snapshot of chromatograph gives you most of the answers. Since both information are missing in your case ---
regards,
FUR
regards,
FUR
-
- Posts: 6
- Joined: Sun Jan 01, 2012 11:08 pm
FUR208 is right
Write more informations about your pneumatic system. What kind of column you use, packed or not? Did you verify if flow rates and temperatures of your unit are proper?
BR
iiYama
Write more informations about your pneumatic system. What kind of column you use, packed or not? Did you verify if flow rates and temperatures of your unit are proper?
BR
iiYama
6 posts
Page 1 of 1
Who is online
In total there are 15 users online :: 1 registered, 0 hidden and 14 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot] and 14 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot] and 14 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science