I am having trouble with a waters 717 autosampler homing the carousel. I wiped down the engaging column with methanol and all the interior to no avail. Looked into the maintenance utilities and the only test that failed was the injector seal calibration (pressure low) which presumably has nothing to do with the homing of the carousel. Any suggestions?
vestel b. shirley, president
betves inc
166 norwood drive
reidsville, nc 27320

Is the instrument not picking up the carousel, or is it picking it up, spinning it and then never finding home? If it picks up and spins, then the carriage drive is fine. It may be a problem with the carriage reader. You want to make sure that the carousel itself is clean, wipe down the top. Make sure that you do not have anything occluding the little holes in the top, and magic marker on it will also cause problems (keep it clean). You could also try another carousel to see if that helps.
2 posts Page 1 of 1

Who is online

In total there are 57 users online :: 2 registered, 0 hidden and 55 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Majestic-12 [Bot] and 55 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry