varian lamp
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
2 posts
Page 1 of 1
hi, I have been using varian 320 uv/vis detector for two years ,yet ,gradully I find that ,the up limit of the absorb value often changs greatly when I change the wave length,and sometims the up limit changes even though I did northing with the detector,i really do not know why is so ?
-
- Posts: 126
- Joined: Mon Aug 15, 2005 4:54 am
Hi
I hope I can help you, there are several possible issues:
First the 320 Detector is running a calibration procedure if you switch it on.
In order to have a good calibration, it is recommended to rinse the flow cell 5 minutes before switching the 320 on ( just start the pumps a little earlier)
This should remove air or sample in the flowcell that might effect the calibration.
Second: There might be some dirt in the flowcell, we did talk about this before, espacially the dual-path cells ( like 9-1 or 4,5-0,15) are sensitive to this issue.
If you have a dual cell, there should be a x on the display
Third: The autorange function might be a possibility, too. You can switch it off
in "setup" --> "integrator/recorder"
Regards Chris
I hope I can help you, there are several possible issues:
First the 320 Detector is running a calibration procedure if you switch it on.
In order to have a good calibration, it is recommended to rinse the flow cell 5 minutes before switching the 320 on ( just start the pumps a little earlier)
This should remove air or sample in the flowcell that might effect the calibration.
Second: There might be some dirt in the flowcell, we did talk about this before, espacially the dual-path cells ( like 9-1 or 4,5-0,15) are sensitive to this issue.
If you have a dual cell, there should be a x on the display
Third: The autorange function might be a possibility, too. You can switch it off
in "setup" --> "integrator/recorder"
Regards Chris
2 posts
Page 1 of 1
Who is online
In total there are 56 users online :: 1 registered, 0 hidden and 55 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Majestic-12 [Bot] and 55 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Majestic-12 [Bot] and 55 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science