We are currently trying to make a calibration curve and our standards consist of ethanol, water, and n-butanol (internal standard). We are operating a 7890B Agilent GC with an ALS. It contains a HP-5 general use capillary column [length=30m , ID= 0.320mm], and has an split/splitless injection.
Our method parameters are:
Injection Volume = 0.5 uL
Inlet Temp = 220 C
Inlet Pressure = 7.5492 psi
Split Ratio = 100:1
Oven Temp = 50-100 C with a 7 C/minute ramp
FID Temp = 250 C
Air Flow = 400 mL/min ; H2 Flow = 30 mL/min ; Make-up Flow = 25 mL/min ; Carrier Gas Flow (He) = 1.5 mL/min
When running our standards we observe a split/shouldered peak for ethanol but the butanol peak is sharp and repeatable. We are unsure of what is causing our ethanol peak to split/shoulder while our butanol peak remains a single sharp peak. When testing 99.5% Ethanol, the shouldering does not occur. Can anyone please help?