Analysis of sugar(Galactose) by HPLC-RID
Posted: Wed Jun 18, 2014 3:46 pm
Dear All,
I have conduct experiment by analysis of 6 sugars which are fructose, glucose, sucrose, maltose, lactose and galactose in general foodstuff.
Can anyone suggest a single column to seperate above six sugar by single HPLC column(if any)?
I currently use the amine column (4.6x150mm, 4um) for anlaysis the five sugar.
For the last sugar(galactose), I use a polymeric sulphonate resin column (8mmx300mm, 7um, Pb Ligand; SP0810)
for analysis.
Actually, the SP0810 column are seldom use.
And, I left the column with mobile phase of water for storage for a year.
After a year, I use the column for analysis on the same standard solution.
Experimental condition:
Injection volumn: 10uL
Flow rate: 500uL/min
Column: Shodex SP-0810
Column temp:80'C
RID detector temp:55'C
refer to the above chromatogram.
I have found that the column seems deteriorated. The peak becomes board peak and tailed with compared with the old chromatogram. The column has not injected any thing for a year...Am I use a wrong mobile phase for storage and therefore deteriorate the column.
I have tried to regenerated the column as the column care manual instructed.
3-4 times 50uL 0.2M lead(II) acetate solution is injected to the column.
However, I cannot restore the column at all.
Can anyone give me advice(s) for this case.
If regeneration is not practical, I shall buy a new column.
Then, how can I keep the column in goods performance apart from using guard column.
By the ways, are there any alternative columns can be used instead when purchase new one.
for analysis.
Column temp:80'C