Good day,
I have a problem , that no peak but it is very small (1% of its true earea and heigh). while the same methods of analysis which I have tried give good response and good results on another HPLC device. I have tried to flush the cell (in its site) with hot water and with water/isopropanol.. no benefit.??
First , check the injector.
If it is an autoinjector , fill a vial with mobile phase and weight it using analytical balance , then make 5 times 100 uL injections and weight again.It should be around 500mg less.

If you are making manual injections , and if it is full loop , make sure you are using a syringe at least 5 times the capacity of your loop.
inject your compound without column and measure peak area. This will tell you if problem is with injector or somewhere else (pump is not delivering right mobile phase and your compound is not eluting).
Vlad Orlovsky
HELIX Chromatography
My opinions might be bias, but I have about 1000 examples to support them. Check our website for new science and applications
www.helixchrom.com
Good day,
I have a problem , that no peak but it is very small (1% of its true earea and heigh). while the same methods of analysis which I have tried give good response and good results on another HPLC device. I have tried to flush the cell (in its site) with hot water and with water/isopropanol.. no benefit.??
Did you try to inject from the same vial on different HPLCs? I am not proud of it, biut I really did make dilution errors by factor 100 reusulting in such a phenomenon :oops:.
Also check if your lamp is working well.
Good day,
I have a problem , that no peak but it is very small (1% of its true earea and heigh). while the same methods of analysis which I have tried give good response and good results on another HPLC device. I have tried to flush the cell (in its site) with hot water and with water/isopropanol.. no benefit.??
Did you try to inject from the same vial on different HPLCs? I am not proud of it, biut I really did make dilution errors by factor 100 reusulting in such a phenomenon :oops:.
Also check if your lamp is working well.
thanks, yes I did it from the same vial on different HPLC.
5 posts Page 1 of 1

Who is online

In total there are 16 users online :: 2 registered, 0 hidden and 14 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot], Semrush [Bot] and 14 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry