I am currently working in a project about organophosphate flame retardants and i have a problem that i would need your help.
Firstly we are using a 7890A GC connected to 5975C MS. Now we are in the stage of running our compounds in Scan mode to find the ions and the retention times of every compound and we are facing a problem with our chromatograms.
In the beginning we were using iso-octane as solution but the cromatograms had an alkyl pattern with many peaks (see foto), and we couldn't recognize the compounds peak. The instrument had installed a CI source and a DB-1 column. We tried many things (clean and change the ion source (install EI source), change the column (install DB-5), clean the detectors and the injector, try in different instrument) and finally we decided to change the solvent. So we made new stock of our compounds, but with toluene as a solvent, and the problem seemed solved. Actually you could still see those weird peaks but with lower abundance!
One day, unfortunately, without changing anything, the peaks are back with big abundance again! The next day the peaks are gone again, but after some runs they appear again! The worst thing is that there is not a standard pattern but it changes by itself, and with that problem we are not able to make a calibration curve! Has anyone anything to suggest or maybe a same experience? The problem maybe caused from the solvent?
I also upload a photo to make myself clear.

I don't know if that is important but by the time that the weird peaks appear the temperature is about 300 oC
Thank you in advance!