New column, worse resolution?
Posted: Tue Jun 10, 2014 11:50 pm
Hello
This pertains to a 6890/5975 setup for running 8270. The new column is the same as the old one: a Zebron ZB-Semivolatiles from Phenomenex. 30m x 0.25 mm x 25 um.
This is the first time I've installed a column... and I'm confused. I conditioned it according to some guide I found somewhere: ran helium through it and baked it out and all that. I ran our curve and almost everything looks good. Retention times are about half a minute later, our later internal standards (Chrysene D12 and Perylene D12) are recovering much better, but mostly everything else looks about the same. However, one thing I don't like is that some pairs of compounds seem to be eluting partly or totally on top of each other - worse than before. p-nitrophenol used to be a shoulder on the right side of dibenzofuran; now they pretty well coelute - depending on the concentration. O-toluidine used to be just touching the peak for m and p cresol; now they are basically merged together.
I thought maybe everything was closer together, but it's not really the case: the time from the first internal to the last is the same as it was before. Plus, other pairs that tend not to fully resolve, like benz[a]anthracene and Chrysene, or benzo b&k flouranthene, look about the same as before.
Anyone have any thoughts?
If it's relavent, here's some more info:
Program held at 40 ºC for 4 minutes
Solvent delay of 3.5 minutes
Increased at 10 ºC/min. to 310 ºC and held for 4 minutes
Column Pressure: 7.1 psi
He Flow:
Column: 1.0mL/min
Velocity: 36 mL/min
Purge: 36mL/min
Total: 40 mL/min
Injector Temperature: 250 ºC
Detector Temperature: 280 ºC
This pertains to a 6890/5975 setup for running 8270. The new column is the same as the old one: a Zebron ZB-Semivolatiles from Phenomenex. 30m x 0.25 mm x 25 um.
This is the first time I've installed a column... and I'm confused. I conditioned it according to some guide I found somewhere: ran helium through it and baked it out and all that. I ran our curve and almost everything looks good. Retention times are about half a minute later, our later internal standards (Chrysene D12 and Perylene D12) are recovering much better, but mostly everything else looks about the same. However, one thing I don't like is that some pairs of compounds seem to be eluting partly or totally on top of each other - worse than before. p-nitrophenol used to be a shoulder on the right side of dibenzofuran; now they pretty well coelute - depending on the concentration. O-toluidine used to be just touching the peak for m and p cresol; now they are basically merged together.
I thought maybe everything was closer together, but it's not really the case: the time from the first internal to the last is the same as it was before. Plus, other pairs that tend not to fully resolve, like benz[a]anthracene and Chrysene, or benzo b&k flouranthene, look about the same as before.
Anyone have any thoughts?
If it's relavent, here's some more info:
Program held at 40 ºC for 4 minutes
Solvent delay of 3.5 minutes
Increased at 10 ºC/min. to 310 ºC and held for 4 minutes
Column Pressure: 7.1 psi
He Flow:
Column: 1.0mL/min
Velocity: 36 mL/min
Purge: 36mL/min
Total: 40 mL/min
Injector Temperature: 250 ºC
Detector Temperature: 280 ºC