I am trying to improve resolutiuon and separation of Endrin and 4,4-DDD. I am using a method shared to me from another lab, whereby they acheive separation on an older model Agilent 6890 GC dual microECD. We have a 7890 GC dual microECD. The peak containing the two analytes appears as a peak with a total width from 19.16-19.32 minutes and at the top of the peak there is a small valley at 19.26 minutes that is about 1/5th in size of the the height of the entire peak.
We are using the following:
Restek liner #23308.5 skiliner splitless double taper gooseneck 4mmX6.5mmX78.5mm
Guard column: Agilent # 160-2535-10 FS, deactivated 0.530X 10m (cut to 5m)
Connector: Restek # 21388 siltek MTX y-connector
Col 1: Restek #11139 30m 0.32IDd 0.5microm df
Col 2: Restek #11324 30m 0.32IDd 0.25microm df
Oven Program:
50 degree C for 0.5 min
50 degree C/min to 100 degree C for 0 min
10 degree C/min to 300 degree C for 4 min
Run time 25.5 min
inject 1 microL
Split/Splitless inlet mode is in pulsed splitless at 210 degree C
pressure 4.9551 psi
total flow 154 mL/min
Septum Purge flow 3 mL/min
injection pulsed pressure 50 psi until 0.5 min purge flow to split vent 15mL/min at 0.4min
average velocity 19.148 cm/sec
holdtime 2.6112 min
Please advise.....
Thank you!