I am having a problem regarding phthalate analysis on a Thermo Focus/Polaris Q ion trap. Last week the chromatography suddenly got really bad, with the peaks showing tailling and reduced sensitivity. I can't figure why this happened, the only dirty samples I injected before this started happening were some headspace VOC's. These were indeed dirty samples, but I thought that headspace analysis doesn't impact the system - only the volatiles are injected.
Anyway, I changed the liner and cleaned the inlet, injected five blanks and then a standard mix at 0.5 ppm. The first standard mix chromatogram looked really good (even if there are some extra peaks - wich can be expected for the first injection after maintenance), but then the peaks started tailling again.
This is an image with three consecutive injection of the same standard:

This is the chromatogram of 0.5 ppm injected on 22.05.2014:

The sample is injected in splitless mode, 1 microliter injection. Injector temperature: 250 C. Solvent: hexane
Oven program: 90 C / 2 min, then 20 C to 275 C for 18 minutes.
The column is inserted in a Thermo Swap it transfer line wich is kept at 310 C. Ion source temp: 250 C.
From the fact that cleaning the injector resulted in a single run improvement, am I right to assume that the problem is with the injector? Does anybody have some advice?
Thanks,
Teodor