Hello,

I am having a problem regarding phthalate analysis on a Thermo Focus/Polaris Q ion trap. Last week the chromatography suddenly got really bad, with the peaks showing tailling and reduced sensitivity. I can't figure why this happened, the only dirty samples I injected before this started happening were some headspace VOC's. These were indeed dirty samples, but I thought that headspace analysis doesn't impact the system - only the volatiles are injected.

Anyway, I changed the liner and cleaned the inlet, injected five blanks and then a standard mix at 0.5 ppm. The first standard mix chromatogram looked really good (even if there are some extra peaks - wich can be expected for the first injection after maintenance), but then the peaks started tailling again.
This is an image with three consecutive injection of the same standard:
Image

This is the chromatogram of 0.5 ppm injected on 22.05.2014:
Image

The sample is injected in splitless mode, 1 microliter injection. Injector temperature: 250 C. Solvent: hexane
Oven program: 90 C / 2 min, then 20 C to 275 C for 18 minutes.
The column is inserted in a Thermo Swap it transfer line wich is kept at 310 C. Ion source temp: 250 C.

From the fact that cleaning the injector resulted in a single run improvement, am I right to assume that the problem is with the injector? Does anybody have some advice?

Thanks,
Teodor
Forgot to add: The colummn is a Thermo TG 5MS, 60m x 0,25 mm x 0,25 um fitted with a 3 m GuardGOLD 0,53 mm by glass union. The flowrate is 1 ml/min.
The splitless time for the inlet is 1 min.
If I am reading the scales in the chromatograms correctly then the tailed peaks are actually taller than the sharp ones, and consequently have much bigger areas.

The only thing that I can think of that would cause this is a split valve that is stuck closed so that you have a permanent splitless injection.

Peter
Peter Apps
Nope, the tailed peaks have a smaller area than the tall ones. The first chromatogram also contains additional peaks (ghost peaks), and so the rest of the peaks are normalized to the tallest peak - in this case one of the ghost peaks.

The peak areas are displayed on the chromatogram with the label AA. My standard mix is composed of six phthalates - the only peaks present in chromatograms two and three.
Does the tailing happen with other analytes as well? If changing the liner did not do the job, I would think of a particle stuck at the head of the column, or something that contaminated the source and causes asdorption on the metal
I haven't tried other analytes, I think I will inject a organochlrine pesticides standard to see how they look.

I'm leaning to think that the inlet is to blame. But what baffles me is that this is the first time I have seen a cleaning restauring performance for one run, and then declining performance again.
Check the septum for tears, and the needle tip for sharp edges. With a damaged needle tip it only takes one injection to dump a piece of rubber into the liner, and non-polar compounds like phthallates just love rubber.

Peter
Peter Apps
I switched the liner again, this time with a new one. The replacement liner that I used earlier was an old one that we washed with several solvents - we don't have a silylating agent, yet. After I installed the new liner, peak shape was as sharp as can be, and tailling didn't start to occur again like with the reconditioned liner.

I guess we will have to start using silylation, as throwing away so many old liners is uneconomical. Thank you for your help.
Teodor
Phthallates are quite tractable compounds, and I doubt that the tailing was dues to interactions with silanols in the inlet, so I would be surprised in silylation would make much difference. The main requirement is to thoroughly clean the liner - general dirt gives non-specific ab/adsorption of compounds like phthallates.

Peter
Peter Apps
9 posts Page 1 of 1

Who is online

In total there are 8 users online :: 2 registered, 0 hidden and 6 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Peter Apps, Semrush [Bot] and 6 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry