EP method for azithromycin

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

6 posts Page 1 of 1
Does anybody run this method successfully? i an dealing with terrible problems. i run in agilent 1200,1260,1290 LC's with XTerra MS. Backpressure is about 90-100bar. After a while, 40-50min(it's 93min gradient!) or maybe in the 2nd or 3rd injection, i have enormous fluctuations, from 90 bar to 30, then to 80 and then to 5 or even 1 bar, done in seconds. if the backpressure doesn't reach 0 bar, the analysis will continue with instant changes from1-10 to 50-70 bar and after a while it may raise up to 80-90 bar. But this will happen again soon and the analysis will stop unavoidably. I have not managed to run more than 7-8 injections in a row. I've used at least 5 LC's and many columns. If someone doesn't know the method and wants to help, he may take a look at EP or ask me for the conditions.
Thanks.
Just a guess without knowing the method: pure acetonitrile as eluent B?
Just a guess without knowing the method: pure acetonitrile as eluent B?
Faulty operation of check valves ? :)

I was convinced that Agilent pumps have active check valves which are not prone to stick.

I would suggest trying different brand of ACN after sonication of dirty valves.
eluent B is ACN/MeOH 75/25. It's a 93min gradient with various steps and a low concentration phosphate buffer for eluent A. i have tried 5 LC's, 2 with active inlet and 3 with the common ball type. i wouldn't blame the instrumentation. HPLCaddict, i really would like to know your thinking behind your answer. Have you seen this before? dblux, we use 3 brands of ACN, well known. .no difference whatsoever. is there someone that really run this method? or recognise the symptoms i described and know what the problem may be ?
eluent B is ACN/MeOH 75/25. It's a 93min gradient with various steps and a low concentration phosphate buffer for eluent A. i have tried 5 LC's, 2 with active inlet and 3 with the common ball type. i wouldn't blame the instrumentation. HPLCaddict, i really would like to know your thinking behind your answer. Have you seen this before? dblux, we use 3 brands of ACN, well known. .no difference whatsoever. is there someone that really run this method? or recognise the symptoms i described and know what the problem may be ?
It's well known problem with some brands of ACN that pumping 100% ACN is imposssible (eg. Dionex LPG 680 pump). However another brand of ACN is OK (Merck LichroSolv gradient type ACN is OK for me even if pumping 100% of ACN).
Eluent B is ACN/MEOH 75/25, not pure ACN. do you believe that using a binary pump i would have better luck?
6 posts Page 1 of 1

Who is online

In total there are 338 users online :: 0 registered, 0 hidden and 338 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 338 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry