I have am having a interesting issue with the D2887 method. I have installed a 2887 compatible column for use with the method, I would condition it per manufacture instructions (40C to 360C at 15C/min, hold for 1-3 hour until baseline is stable) and it would appear fine until I start running samples.
The first baseline would be perfect, and any baselines after that would be fine, I run the calibrant next and the peaks are nicely resolved and calibration takes, I would next run RGO 1 and the chromatography would appear okay, but everytime I run a baseline afterward the baseline would rise over 100pA when approaching the higher temp range of the method.
I have repeated this phenomenon with three different columns and at a loss as how to fix it.
I think I ruled out column damage because the baseline only rises at higher temperature (300C +) and the 40 to 200C baseline looks fine and 4 of us have swept through all the plumbing to rule out leaks.. Could be column contamination but then I am not sure how to fix it given the analytes. Could be incomplete conditioning but it looked complete when it was conditioned and this only occurs after a calibrant and rgo is analyzed.
Any ideas would be greatly appreciated