Replicates of calibration curve lowest standard

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

4 posts Page 1 of 1
Hi everyone,

I have a question regarding calibration curve lowest standard. Do you prepare lowest and highest standards in duplicate?
In this case are you using mean value or the first preparation value in the regression?

Thanks in advance! :)
nope, we don't do this.
----suffers separation anxiety----
In principle, you should use *all* the data points.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
Depends on what work you are doing. If you are running under EPA methods it isn't allowed, you only use one standard determination per level, no averaging. I do often prepare them in duplicate and use them to QC the curve afterwards, since if the curve is good and instrument is behaving properly you should be able to quantify them back accurately. Lowest should be able to hit +/- 50% and the high +/- 20% or much less.
The past is there to guide us into the future, not to dwell in.
4 posts Page 1 of 1

Who is online

In total there are 55 users online :: 0 registered, 0 hidden and 55 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 55 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry