Analyst 1.6.2 and Flexar FX-15 compatibility on Windows 7?

Posted: Mon May 26, 2014 3:08 pm
by koeweg
Just wondering:

Who has Analyst 1.6.2 running on a Windows 7 (32bits) system and is working with a Perkin Elmer Flexar FX-15 UHPLC system?
We're trying to upgrade from Analyst 1.4.2 to 1.6.2 and from WinXP to Win7, but can't manage to get it to work together..
When I run a sample it seems like the system doesn't get a stop code to stop the run at the end of a run, it just keeps on going until it reaches the idle time limit for the pumps.
I can't find any driver updates for Perkin Elmer Flexar FX-15.
For Analyst 1.6.2 I found some fixes, but not for my problem..

Regards,
J.

Re: Analyst 1.6.2 and Flexar FX-15 compatibility on Windows

Posted: Fri Jun 06, 2014 6:25 am
by emeulman
Analyst only supports the PE 200 series instruments. I guess you have setup your hardware profile in Analyst as MS only where you only trigger the MS when it gets a start or stop signal from your UHPLC. When your run is done, the MS will to to standby, but your system will keep op pumping. Can't you add a second method in your queue with a switch, where you lower your flow and switch the flow to waste instead of MS?

Re: Analyst 1.6.2 and Flexar FX-15 compatibility on Windows

Posted: Fri Jun 06, 2014 7:02 am
by koeweg
Hi,

I think I fixed it. Turned out that the newest Flexar FX-15 pump driver (ver. 1.5) was the problem.
Downgrading to version 1.0.1.xx solved the problem. :)

Thanks for the suggestion!

Re: Analyst 1.6.2 and Flexar FX-15 compatibility on Windows 7?

Posted: Mon Mar 28, 2016 11:07 am
by alexkirk
Hi All,

Could anyone send me the copy of Flexar driver for Analyst (version 1.5)?
We want to it out. We have only older version.

Regards,

A

Re: Analyst 1.6.2 and Flexar FX-15 compatibility on Windows 7?

Posted: Wed Mar 08, 2017 7:25 pm
by Frogless
I'm trying to connect a Flexar pump to any version of Analyst. I need a copy of the Analyst drivers for the Flexar.

Does anyone have a copy or point me to a download?

Re: Analyst 1.6.2 and Flexar FX-15 compatibility on Windows 7?

Posted: Wed Mar 08, 2017 8:06 pm
by Frogless
I actually found a copy of the Analyst driver for the new Flexar FX-15 and FX-20 LC systems. Message me if anyone needs a copy.