We have a new IC system ICS-2000 from Dionex, which was modified to use external water to regenerate suppressor. The analyte is bethanechol Cl and we follow USP method exactly. The detailed method is:
MP: 29mg EDTA and 0.3 mL HNO3 in 1 L water. 950 mL the solution mixed with 50 mL of ACN.
Column: Waters IC-Pak Cation M/D, 3.9x150 mm
Cell temp: 35C
Column temp: 30 C
Suppressor: CSRS-Ultra II at 15 mA (for cation)
Sample and standard prepared in 0.1 N HCl
Bethanechol elute around 11 minutes
We had no problem running samples using the system until recently. The column appears degrading pretty fast and at certain point we saw the main peak getting a shoulder (same to other peaks from sodium, etc). After I repaced with a new column, we were surprised to see a well-defined negative peak at same retention time (11 min) but no bethanechol peak. The area count slightly smaller with similar peak shape. Sodium peak in the front is huge and positive as usual.
Have tried:
1. Injected known-good sample and standard;
2. Used known-good MP;
3. Flushed the system with water;
4. Flushed the suppressor with 40ACN/60water/1Methanesulfonic acid;
5. Put back the bad column, and positive bethanechol peak no longer there;
6. Inject 0.1 N HCl, and no peak at 11 min;
7. Replace the external water for supressor regeneration.
The only thing i am not sure is the working condition of suppressor, but even a weak suppressor shouldn't produce a negative peak at 11 min. When the system was good, total conductivity was about 0.6 uS and now it is about 1.5 uS. Where did the additional 0.9 uS come (if the reading means something)? When I turned off suppressor I can see base line gradually moving up, indicating it is at least working. There is nothing I can think of except getting the service tech coming to my place with a new suppressor. But they were quite confused on the phone.
A negative peak means conductivity at that retention time is lower than the back ground, which is not necessarily unusual, but how could that be the negative peak elutes at 11 minutes? What else did I miss?
Having working on LC for many years, I got really frustrated for the first time. Any suggestion is very welcome. Hope the new suppressor will solve my mystery.