Automate my manual events integration in Chemstation B.03.01
Posted: Fri May 16, 2014 5:21 pm
Hello all.
I would like to automate my manual events integration On a 1200 LC using ChemStation B03.01. Let me explain further.
We are a food ingredients R&D lab that analyze a lot of proteins and hydrolysates by Gel permeation SEC. We then need to integrate the chromatogram based on time and not by peaks because we need to know the % within a certain molecular weight range. We run proteins and peptides of know MW on the column and "calibrate" the column. In the past with our 1050 we would bring up a chromatogram in Data analysis, delete all of the peaks, manually redraw the baseline, and manually split the peaks at specific times correlating to the MW we are interested in. We would then click "Copy Manual Events to Method" and make sure the Manual Events box is checked in the events table.
It seems to me that we should be able to do this in the timed events table. I can get a good baseline drawn, and have entered "split peaks" at the times we want. However, the resulting chromatogram does not integrate according to these times. I entered split peaks at 5.968, 6.591, 7.214, 8.036, 8.659, 9.282, & 10.727 minutes. ChemStation then integrates at 6.522, 7.143, 7.765, 8.386, 9.007, 9.628, 9.628 (yes twice), & 10.871 minutes.
Does anyone have any idea how I can automate this integration, it would be so much easier that doing it manually.
Thanks.
Thanks.