by
varossf » Tue May 20, 2014 8:23 pm
HI,
Advanced reporting is quite simple and there is a good manual that can help you out.
EX.R = repeating formula
PEAK.ESTDCONCENTRATION = concentration ESTD corrected for sample amount dilutors multipliers
RA = All rows in the seqeunce (A you can replace with e.g. specific row number
Row Number: the number of the sample that is included in the advanced report, not necessarily the sequence row number
1: downwards
T1: trace 1 like DAD-A
PA: all peaks, but you can replace the A for selecting a specific peak from the peak table
3: Identified and unidentified peaks
Freek
Hi ,
Anybody using advanced reports for Chromquest/Open Lab?
I cant seem to figure out how to link these with the sequences and struggling to figure out what the formulas actually mean
ie
=EX.R (PEAK.ESTDCONCENTRATION(),"RA;1;0,"T1","PA;3;0;0").
The manual is very vague on this reporting
My background is Chromeleon, Empower, chemstation , Analyst so thought this would be straightforward,, but I'm slowly losing the will to live with this software
Thanks