we are seeing higher response for a particular compound (low level 0.3 ug/ml) on brand new column. As the column is getting older the response drops to about half, but the peak shape remains pretty much the same - nice symmetrical

anybody any thoughts as to what might be going on?

thanks
Some questions:
  • Are you measuring peak area or peak height?
    By "response", do you mean the slope of the calibration curve (the "response factor") or are you referring to the size of the peak at a single level?
    Is the plate count staying constant? If not, how much is it changing?
    Any change in column pressure?
    Any change in retention time?
    Has the decrease in response been gradual, or did it occur abruptly?
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
In addition to Tom my questions are:
- what do you mean with old - 1 day, one week, one month?
- how often you use a fresh standard
- do you see any degradation product?
- UV detection?
Gerhard Kratz, Kratz_Gerhard@web.de
3 posts Page 1 of 1

Who is online

In total there are 10 users online :: 0 registered, 0 hidden and 10 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 10 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry