Multiple values per level in ChemStation calibration table
Posted: Tue May 13, 2014 5:38 pm
by koenvanderdrift
Is it possible to use more than one value per level in a calibration table? I found the option to either average or replace values, but not to add. We're using version B.03.01.
Re: Multiple values per level in ChemStation calibration tab
Posted: Tue May 13, 2014 6:40 pm
by dblux_
To have more calibration points with the same concentrations simply add them as consecutive levels (eg. Lvl 1 - conc 0.1; Lvl - 2 conc. 0.1; Lvl 3 - conc. 0.6; Lvl 4 - conc. 0.6 and so on).
Re: Multiple values per level in ChemStation calibration tab
Posted: Fri May 16, 2014 1:19 pm
by koenvanderdrift
Thanks!