Autosampler issues
Discussions about GC and other "gas phase" separation techniques.
3 posts
Page 1 of 1
I have an HP 7673A autosampler connected to a 5890 that is having some trouble. It cannot not pick the vial back up after the injection has been made. It seems that the arm doesn't line up properly with the vial. It is off by about a 1/4 inch. It tries to grab it but hits the neck. The arm resets and tries one more time then gives an error and stops the run. It doesn't have a problem placing the vial. I've tried to see if the vial holder is sticking or isn't homing properly but I feel like if that were the case, the syringe would jam. Any help is greatly appreciated.
-
- Posts: 3594
- Joined: Mon Aug 30, 2004 11:43 pm
Can you post a photo (like use photobucket) here? My first guess is that one of the rods is sticking because of dirt, and isn't spring back into place.
-
- Posts: 350
- Joined: Tue Jun 28, 2011 7:30 pm
This one has usually stumped me. As CPG says, cleaning up the rods under the gripper arm might help; the plastic band inside the arm may be replaced as well.
I have not had any luck with these measures - I usually end up sending the tray to GenTech.
I have not had any luck with these measures - I usually end up sending the tray to GenTech.
3 posts
Page 1 of 1
Who is online
In total there are 170 users online :: 0 registered, 0 hidden and 170 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 170 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 170 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science