XCalibur - Activate a stand by mode at the end of a sequence

Posted: Sat May 10, 2014 9:04 am
by klam
Hi,

As you will understand I'm not an expert...

So I have a Gc Trace Focus - MS ITQ 700 and I would like to decrease save my helium when the sequence of analysis is done.

How can I tell the software to decrease the temperature of the GC to 30°C and stop the flow of helium?

Thanks a lot,

Kevin

Re: XCalibur - Activate a stand by mode at the end of a sequ

Posted: Mon May 12, 2014 9:37 am
by BMU_VMW
we simply add one line at the end of the sequence with an instrument methode that has a x-fer line temp of 110°C, GC-oven of 40°C and a reduced He flow. simply inject a blank solvent or air to make sure the autosampler won't cause an error.

Re: XCalibur - Activate a stand by mode at the end of a sequ

Posted: Mon May 12, 2014 5:37 pm
by ThermoMarkB
Hi Kevin,

I would NEVER turn off helium low (or carrier in general), this is one of the worst things you can do to a GC or GCMS instrument. I ould go into a littany of reasons why, but believe me, it's bad.

Thermo uses the Gas Saver option in the method to turn down the total flow after the injection is made, you can usually find this option in the Carrier or Inlet setup tabs of the instrument method. By default the hardware will be in this "saving" state until it gets a trigger to prepare for the next injections, so relax, you don't really have to do anything, you're already saving gas. If, however, you really want to maximize the savings for a long-term standby, simply set the column flow to ~0.5mL/min and leave the inlet temp at 225°C. The inlets and detectors are much more stable and clean when left hot and inert. Same thing for the mass specs, just leave the transfer line hot and the ion source hot and obviously keep the MS on. I promise you, the instruments will be much more stable, clean and healthy doing this rather than a shutdown.

-Mark

Re: XCalibur - Activate a stand by mode at the end of a sequ

Posted: Mon May 12, 2014 11:48 pm
by klam
Thanks!

Would a splitless and 0.5ml / min be ok ?