do you work at high senitivity?(analysis at the ppb range)
Advertisement
Discussions about GC and other "gas phase" separation techniques.
helloYour injector or detector temperature maybe too low.
Your column temperature may not focus your injection plug properly.
Your injection technique may not be suitable.
Your detector may have the wrong flows. Check with a known std for sensitivity of the detector.
Your actual split ratio may not be what you think it is.
good luck.
thanks for replyHi again, you must detect your sample at 1 ppm. I suppose you're using the 10^3 range, and working with a FID. Are the H2 and air flows in the optimum range?
Anyway, your can work with a low split ratio, i.e. 1:10.
BTW what is your sample's solvent?