Page 1 of 1

How to program Empower to make autosample to mix/inj?

Posted: Tue Apr 29, 2014 4:41 pm
by tpmun
Hello there,

We are trying to run a new method which requires the autosampler to mix 15
ul of common reagent with 15 ul of sample for 1 min before inject this mixture
into the HPLC column.
Would you be able to give us any suggestion on how to program Empower 1 to
do it? :?

Thank you.

Re: How to program Empower to make autosample to mix/inj?

Posted: Wed Apr 30, 2014 2:33 am
by DR
What model LC are you using??

Re: How to program Empower to make autosample to mix/inj?

Posted: Wed Apr 30, 2014 10:10 am
by aceto_81
If you are using an Alliance, you can do this with autoadditions.


Ace

Re: How to program Empower to make autosample to mix/inj?

Posted: Wed Apr 30, 2014 6:41 pm
by tpmun
Thank you for your messages.

We have Waters 717plus Autosampler, Waters 600E Multisolvent Delivery System and Shimadzu RF-535 Fluorescence Monitor. They are controlled by Empower 1 software.

Recently, we be able to program the autosample to mix reagents and then inject the mixture. Unfortunately, we have no idea how to make the Empower controls the autosampler to do that trick.