Response of Ar v O2 on a TCD

Discussions about GC and other "gas phase" separation techniques.

4 posts Page 1 of 1
In theory, Ar should give a larger response (per unit vol) than oxygen with He carrier gas (due to having higher TC differential from He than oxygen), but practically, I've found that this may not be the case. Can anyone explain please?

Thanks...

A few factors affect the sensitivity of TCD. In addition to thermal conductivity difference between the carrier gas and the analyte, molecular weight of the analyte is another one. Oxygen has a smaller molecular weight which gives it better response. Also, the sensitivity of TCD is very sensitive to the temp of the detector. The sensitivity goes down as TCD temp increases. Argon and oxygen could behave differently in different TCD temp.

Thanks for the reply, it explains a few things. Much appreciated.

The word ..may... doesn't tell us much, however, I was told TCD's MAJOR PROBLEM was keeping the dual flows of detector reference gas constant.
4 posts Page 1 of 1

Who is online

In total there are 51 users online :: 0 registered, 0 hidden and 51 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 51 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry