Low intensity in TSQ Vantage

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

4 posts Page 1 of 1
Hi everybody,
I am working with a LC-MS/MS TSQ Vantage HESI II. The instrument has been running properly during last 5 years but suddenly I faced a significant decrease in intensity and the system cannot get calibrated using polytyrosine due to low signal. Can anybody give me some information where the problem is?
Thank you in advance.
First thing I would do is do a clean of the front end optics.. Up to and including Exit lens on API stack.. Be very thorough on the S-lens and the inside of the orifices on the lenses.. Use MicroMesh 6000 or similar.

Also verify the metal needle in the Probe is not completely blocked or burned at the tip.

Good luck.
Kind regards
Leadazide
Dear leadazide,

Thank you very much for your invaluable information. I did as you had suggested and now the instrument is working properly.

Best regards
Glad to help! Feel free to ask again if you run into other issues.
Kind regards
Leadazide
4 posts Page 1 of 1

Who is online

In total there are 15 users online :: 1 registered, 0 hidden and 14 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Amazon [Bot] and 14 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry