- Posts: 762
- Joined: Thu Dec 13, 2007 5:54 am
I'm injecting 10uL (in microliter pickup mode with my Varian ProStar 410 autosampler, with DI water as my co-injected solvent) of 96:4 Methanol:DI Water out of a polypropylene vial with an aluminum crimp cap and polyethylene septa onto a Waters dC18 2.1x150mm 3um column. I'm running 5mM Sigma LCMS grade ammonium acetate in DI water (filtered through a 0.2um GHP membrane filter) on pump B, and 100% Macron LC/MS methanol on pump A (both pumps are Varian ProStar 210s), running 0.25mL/min at 45C. Gradient is as follows: 5:95 A:B at t0, linear gradient to 100:0 A:B at 40 min, hold 100:0 to 60 min.
I'm running in profile mode, scanning 50-450 m/z. Parameters are as follows: Positive - needle at 5000V, shield at 600V, negative - needle = -4500V, shield = -600V, drying gas temperature = 400C, nebulizing gas = 56psi, drying gas =36psi, CID gas = 2 mTorr. detector at 1800V.
Positive TIC

Negative TIC

Negative spectrum at ~40min, roughly the maxima of the negative TIC.

Positive spectrum at ~40 min

Positive ion spectrum of peak in positive TIC at 31.1 min