Hi all,
I got a old Knauer c18 HPLC column from my colleague ... I could not ask him how and what for he has used the column
The CoA of the column says @1ml/min 80/20 MeOH/water the back pressure would be around 170bar ..
but I get ~240bar ... I first want to try cleaning the column ...
Can some one give me a step by step method of reviving the column ... a nice link might also suffice ...

This is the first time I am trying to do something of this sort ...


with best wishes
HPLC System: HPG system with two(2) Knauer S1000 pumps- Dyamic Mixing Chamber-Knauer s3900 auto sampler-S2300 RI detector-S2500 UV detector- Raytest radioactivity detector

tarzan
First of all take the column out of your system, open the column top and replace the frit!
Than built in the column in reversed flow but don't connect it to the detector.
Reduce flow and flush first with water, than with organic solvents like Isopropanol, Methanol and/or Acetonitrile.
You can also flush with "hot" water.
You can use the column after this cleaning procedure in reversed flow direction. Backpressure should be lower than 240bar but will not go down to 170bar. But you can work with the column
Good luck
Gerhard Kratz, Kratz_Gerhard@web.de
thanks dear Kratz ..
Actually I just ran benzene sample .. as it was used in the column CoA ...
The retention time given in CoA is 4.8min and I get 4.5 min ... of couse as i said the back pressure is 240 bar ..
but the peak has a bit of tailing ...
so what should i do .. can i use the column .. or i should still go for cleaning it ..?
HPLC System: HPG system with two(2) Knauer S1000 pumps- Dyamic Mixing Chamber-Knauer s3900 auto sampler-S2300 RI detector-S2500 UV detector- Raytest radioactivity detector

tarzan
I would do the cleaning following this protocol:
Regeneration of RP packings
C18, C8, C4, C1, C30, CN and Phenyl stationary phases:
• Flush the column with 20 column volumes of water
• Flush the column with 20 column volumes of acetonitrile
• Flush the column with 5 column volumes of isopropanol
• Flush the column with 20 column volumes of heptane
• Flush the column with 5 column volumes of isopropanol
• Flush the column with 20 column volumes of acetonitrile

In addition I would run the column in reversed flow direction.
Good luck.
Gerhard Kratz, Kratz_Gerhard@web.de
.... but don't spend an awful lot of time and solvent on an old column. If you look at what your time is worth, and the amount you have spent on your samples already, and balance this against the price of a new column, you may find that columns are more consumable than they appear at first sight!

If you can get 500 injections out of a column, it has almost certainly cost you less than the plastic-ware and vial you used to make the samples (a spin-filter is nearly a pound in the UK, good quality vial, insert and cap is getting on for a pound, and a typical hplc column maybe 500 pounds).
5 posts Page 1 of 1

Who is online

In total there are 7 users online :: 1 registered, 0 hidden and 6 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Semrush [Bot] and 6 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry