Can anyone tell me what is the difference between mass gain and mass offset versus AMU gain and AMU offset?
----suffers separation anxiety----
Amu gain and offset are parameters that affect the mass peak width, while mass gain and offset are parameters that affect the mass axis calibration.
Amu gain and offset are parameters that affect the mass peak width, while mass gain and offset are parameters that affect the mass axis calibration.
I believe the Gain changes the high mass faster than the low mass and the Offset changes them at equal amounts. Of course I could have that backwards but that is the difference between gain and offset.
The past is there to guide us into the future, not to dwell in.
Thanks for the replies.

I understand the difference between gain and offset - as James stated, on the Mathieu stability diagram, gain increases the slope of the scan line and offset increases the y-axis intercept.

I see how this affects peak width and resolution, but I'm not sure what the mass axis calibration is all about
----suffers separation anxiety----
The mass axis calibration converts your experimentally measured intensity-time signal into intensity-m/z plot in order to obtain an accurate mass determination of the ions in the mass spec.
Thanks for the replies.

I understand the difference between gain and offset - as James stated, on the Mathieu stability diagram, gain increases the slope of the scan line and offset increases the y-axis intercept.

I see how this affects peak width and resolution, but I'm not sure what the mass axis calibration is all about
It makes an ion at 218.6m/z move to 219.0 m/z to correct the mass assignment.
The past is there to guide us into the future, not to dwell in.
Thanks James and Jerole - I think I get it now

So increasing the gain of the mass axis calibration will 'stretch' the m/z axis to fit the time axis optimally, and offset will shunt it along.
----suffers separation anxiety----
7 posts Page 1 of 1

Who is online

In total there are 18 users online :: 0 registered, 0 hidden and 18 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 18 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry