Any suggestions about why I might have a large peak in the void volume of my LC/MS/MS runs?
Do I have overload and therefore this analyte is not retained?

Need more information in order to comment:
- Are you running a pure standard or an actual sample?
- Are you looking at the total ion chromatogram or a specific ion?
- Do you have a big peak at t0 and then your analyte peak, or just one huge peak starting at t0?
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374

this was actual samples (extracts from biota)
and it was the specific transition (from MS/MS) that had the strange peaks
And there are two peaks, the first one at t0 and the second one at the normal retention time.

Please see the note that I just posted to Ericz. It could be that your sample is dissolved in an organic solvent, and your sample migrates through the column unretained.

Or you got still so much junk in your sample that you see every mass in your unretained peak.
5 posts Page 1 of 1

Who is online

In total there are 34 users online :: 3 registered, 0 hidden and 31 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Bing [Bot], Semrush [Bot] and 31 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry