Varian 220MS problem
Posted: Mon Apr 14, 2014 4:50 pm
Hi Everyone
My Varian 220MS is showing a high (fluctuating between OK and High) cal gas value even when I completely close the cal gas valve.
I am not seeing any peaks for the cal gas or air and water.
The diagnostics pass without any problems and the filaments are not open.
The RF looks fine.
The turbo is 100% (230 mA).
I have cleaned the trap thinking it could be contamination but this did not help.
Any other ideas would be warmly received.
Thanks in advance.