decreasing the Peak area in sample For Agilent 1200 series
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
6 posts
Page 1 of 1
How can i know the possible reasons for decreasing the Peak area in the sample , i have tried all the possibilities ,Pl help
-
- tom jupille
- Site Admin
-
- Posts: 4978
- Joined: Wed Aug 11, 2004 4:55 pm
Obviously you have not tried *all* the possibilities, since the problem is still there. So, why don't you tell us what you *have* tried and we can go from there.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
-
- Posts: 3594
- Joined: Mon Aug 30, 2004 11:43 pm
I can't even understand if you WANT to decrease the peak area (like using a different wavelength not at peak apex), or if you're observing peak area decrease over time.How can i know the possible reasons for decreasing the Peak area in the sample , i have tried all the possibilities ,Pl help
Everyone needs to provide details. And most stuff is not nearly as "proprietary or secret" as one thinks.
-
- Posts: 4
- Joined: Sat Apr 12, 2014 9:32 am
Hi ,
i have tried
1.replacing the standards with new standards
2.washed the column
3.clean & sonicated the needle seat
4.checked the Lamp health
i have tried
1.replacing the standards with new standards
2.washed the column
3.clean & sonicated the needle seat
4.checked the Lamp health
-
- Posts: 38
- Joined: Mon Sep 13, 2010 8:46 am
What is the sampler, manual or automatic, which model? Is there pressure drop when the area decrease?
-
- Posts: 686
- Joined: Mon Apr 29, 2013 4:37 pm
Hi ranjithmuthoju,
Here's a list of some possibilities (not an exhaustive list)...really, though, more information about what you are trying to separate (that is, for what type of compound is the peak area decreasing...what type of detection and the like) would be helpful. Please, see what you think...some I list are somewhat obvious, perhaps.
1. Decrease in injection volume, or perhaps, a leak in the system between the injector and column.
2. Increase in eluent flow rate, or perhaps a leak that occurred between the column and detector somehow "got fixed".
3. If UV-type detection, a change in wavelength.
4. A change in pH value of the eluent (either with or without the sample dissolved in it).
5. Adsorption of the analyte somewhere within the chromatograph.
6. Solvent mismatch between the sample and the eluent...in particular with early-eluting analytes.
Here's a list of some possibilities (not an exhaustive list)...really, though, more information about what you are trying to separate (that is, for what type of compound is the peak area decreasing...what type of detection and the like) would be helpful. Please, see what you think...some I list are somewhat obvious, perhaps.
1. Decrease in injection volume, or perhaps, a leak in the system between the injector and column.
2. Increase in eluent flow rate, or perhaps a leak that occurred between the column and detector somehow "got fixed".
3. If UV-type detection, a change in wavelength.
4. A change in pH value of the eluent (either with or without the sample dissolved in it).
5. Adsorption of the analyte somewhere within the chromatograph.
6. Solvent mismatch between the sample and the eluent...in particular with early-eluting analytes.
MattM
6 posts
Page 1 of 1
Who is online
In total there are 12 users online :: 2 registered, 0 hidden and 10 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Baidu [Spider], Google [Bot] and 10 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Baidu [Spider], Google [Bot] and 10 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science