Conditioning column
Posted: Wed Apr 09, 2014 8:52 pm
by MaryL14
I installed a guard column with my analytical column for the first time. Guard column on inlet end.
Checked for leaks - none. Helium through column present (placed free end in methanol to check).
Conditioned column for 8 hours
Installed into MS, ran blanks. baseline rise, column bleed present.
Leak? Although, I did not detect it prior to conditioning column.
With that long of column conditioning (8 hours), if there was a leak present, the integrity of the column is questionable? Is there a chance that the contamination was kept to just the head of the column.
Troubleshooting since column bleed: Approximately 2 meters were trimmed from the head of the analytical column.
Re: Conditioning column
Posted: Wed Apr 09, 2014 11:01 pm
by Consumer Products Guy
I typically haven't conditioned my new capillary columns for years now. Reason is that the columns are all pre-run by manufacturer inside a GC (chromatogram is provided with each column, so it has already gone through a GC programming cycle).
Remember - not everything I do is correct.
Posted: Thu Apr 10, 2014 1:26 pm
by Peter Apps
It is not clear whether the column was new - if it was a used column then the "bleed" could be heavy contaminants slowly eluting.
Posted: Sat Apr 12, 2014 8:27 am
by uzman
Did you check the ferrules , they must be graphite/vespel types for MS , not pure graphite.
Posted: Sun Apr 13, 2014 2:04 am
by Don_Hilton
Type and age of column of column? What was the condition of the column before installation?
What was the treatment of the guard column?
How did you connect them?
Temperature range ramp and used for conditioning?
What grade of carrier gas are you using, do you have an indicating trap on the instrument?
What was the head pressure of the instrument throught the conditioning run and were you running split or splitless?
A column in unknown condition leaves us no information as to wether a change has taken place.
The other questions relate to some things that can be done to damage a column - particularly taking the column over temperature or inadvertant introduction of air at high temperture over a long period of time.
Posted: Mon Apr 14, 2014 7:46 pm
by osp001
Not OP, but I did not know this.
Why not straight graphite?
Posted: Mon Apr 14, 2014 9:57 pm
by dblux_
"The 15% graphite/85% Vespel ferrules are
used at the MS interface as they are non-porous to
oxygen, thus prohibiting oxygen from entering the MS."
Posted: Wed Apr 16, 2014 10:56 pm
by Don_Hilton
The big problem with 100% graphite is that they are very soft and will creep - with loss of the seal between the column and the fitting. You need to tighten ferrules ocasionally - graphite more often and after a while you have extruded all the graphte to wherever it goes. With a a mass spec, you have some risk of graphite fragments being sucked into the instrument and interfering with the electronics.
Posted: Sun Apr 27, 2014 11:39 pm
by mrussell
Agilent recomends against conditioning a colunmn from more than two hours.